Posted: Wed Sep 12, 2012 7:58 pm
Two months ago I prepared solutions of benzene in methanol at different concentrations ranging from 25mg/L up to 400mg/L and I used ethyl acetate in methanol at 500mg/L as my internal standard. Back then I did 3 injections of each benzene solution with my internal standard to get a calibration curve with the ratio of area of benzene peak to area of ethyl acetate peak as my y-axis and benzene concentration as my x-axis. The curve had a good linear fitting with R^2=0.993.
I had to make the same calibration curve a couple of days ago and I prepared fresh solutions of benzene and ethyl acetate and ran the injections, except I did two injections this time per solution rather than three. All the GC settings are the same, I used the same chemicals as before and also the amount of chemicals I used was the same as before. However, this time the calibration curve isn't as good this time around, the R^2 value is lower and it's 0.971.
What caused the change in the calibration curve fitting, how can I check what is the culprit?
My second part of the question is: what is the typical procedure when making internal standard calibrations in terms of number of injections? And is it proper to use my calibration curve from 2 months back?
Thank you.
I had to make the same calibration curve a couple of days ago and I prepared fresh solutions of benzene and ethyl acetate and ran the injections, except I did two injections this time per solution rather than three. All the GC settings are the same, I used the same chemicals as before and also the amount of chemicals I used was the same as before. However, this time the calibration curve isn't as good this time around, the R^2 value is lower and it's 0.971.
What caused the change in the calibration curve fitting, how can I check what is the culprit?
My second part of the question is: what is the typical procedure when making internal standard calibrations in terms of number of injections? And is it proper to use my calibration curve from 2 months back?
Thank you.