Column backpressure
Posted: Wed Sep 12, 2012 2:21 pm
				
				How can I find that the high backpressure of the column (c18) is related to packing deterioration or contamintion? in the other hand, how I can check the integrity of column packing?
			Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE).  If you are a registered user, please log in.
http://www.chromforum.org/
Before blaming the column, however, it's a good idea to check for other potential blockages. I like to disconnect fittings starting at the end of the system and working my way upstream. You may find the blockage is at an in-line filter (you *are* using and in-line filter, aren't you?), or a transfer line, or the injection valve.How can I find that the high backpressure of the column (c18) is related to packing deterioration or contamintion?