GC-MS to GC area conversion

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Im sorry if this query was already posted before. I tried searching through previous post but couldn't find anything regarding this. When i run samples in GC-MS, the area for chloroform that I obtained was 1,971,183,624. But when the samples were analyzed in GC, the area (pA*S) that i got was 341,144. Is there any way to convert the area that I obtained fr GC-MS to the area in GC or vice-versa?? What's the unit of area in GC-MS? TQVM for anyone who answers my question.
Response factors vary from detector to detector. The area for chloroform will vary in the mass spec depending on wether you sue the signal for one ion, another ion, or all ions (in the mass range acquired).

The best solution - build a calibration curve for each instrument.

There have been discussions of mass spec response relative to TIC on this site.
hey... thanx so much for the reply. Im doing the same thing as u suggested now. separate calibration curve for GC and GC-MS...
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry