Normal Phase Problems
Posted: Tue Sep 11, 2012 10:46 pm
First, thanks for all the information I've found on this site while just lurking. You've all taught me a great deal is I've grown in experience over the years. Unfortunately, I'm haveing some trouble right now that I just can't seem to get past and was hoping someone here would have some insight. I'm attempting to PQ an Agilent 1200SL with the following modules:
Degasser - G1379B
Pump - G1312B - Binary 600 bar
Autosampler - G1329B - Standard
Column Compartment - G1316B
Detector - G1315C - Diode Array
We have several identical/similar systems, most of which are reverse phase and give us no problems.
Two of the systems are set up with normal phase analyses (essentially hexanes for mobile phases), one of which is gradient. It's these two systems that give me headaches. The first system that started acting up was the isocratic assay, which started displaying some excessive pressure fluctuations and an unstable baseline with peak shifting after a few months. Eventually, I ended up cleaning the system out with dilute nitric acid and DMSO (using IPA and/or acetone as a transition solvent) and the problem went away.
Currently, I'm attempting to PQ the gradient assay system, and seeing what appears to be the same issue. The signal noise and drift test uses hexanes @ 1mL/min for the mobile phase, 254nm wavelength, no injection and a restriction capillary to add some backpressure (~20 bar) and delay volume (~0.5mL).
I'm getting a total pressure fluctuation of ~0.5 bar that doesn't seem to have any particular pattern. Normally, I would expect minimal pressure fluctuation (0.1 bar, if any). I'm also getting somewhat excessive noise (~5x what I would expect) and baseline drift.
The things I've tried:
Clean system with dilute nitric acid (5%) and DMSO (each at least a couple of hours @ low flow), using IPA as transition solvent.
Change pump seals (no unusual wear seen) and purge valve frit.
Remove the damper and mixer from the flow path.
Plumb as an isocratic system.
Bypass sampler.
Nothing seems to be working. Any insight anyone could provide would be greatly appreciated.
Thank you,
Eric
Degasser - G1379B
Pump - G1312B - Binary 600 bar
Autosampler - G1329B - Standard
Column Compartment - G1316B
Detector - G1315C - Diode Array
We have several identical/similar systems, most of which are reverse phase and give us no problems.
Two of the systems are set up with normal phase analyses (essentially hexanes for mobile phases), one of which is gradient. It's these two systems that give me headaches. The first system that started acting up was the isocratic assay, which started displaying some excessive pressure fluctuations and an unstable baseline with peak shifting after a few months. Eventually, I ended up cleaning the system out with dilute nitric acid and DMSO (using IPA and/or acetone as a transition solvent) and the problem went away.
Currently, I'm attempting to PQ the gradient assay system, and seeing what appears to be the same issue. The signal noise and drift test uses hexanes @ 1mL/min for the mobile phase, 254nm wavelength, no injection and a restriction capillary to add some backpressure (~20 bar) and delay volume (~0.5mL).
I'm getting a total pressure fluctuation of ~0.5 bar that doesn't seem to have any particular pattern. Normally, I would expect minimal pressure fluctuation (0.1 bar, if any). I'm also getting somewhat excessive noise (~5x what I would expect) and baseline drift.
The things I've tried:
Clean system with dilute nitric acid (5%) and DMSO (each at least a couple of hours @ low flow), using IPA as transition solvent.
Change pump seals (no unusual wear seen) and purge valve frit.
Remove the damper and mixer from the flow path.
Plumb as an isocratic system.
Bypass sampler.
Nothing seems to be working. Any insight anyone could provide would be greatly appreciated.