Advertisement

HPLC Carryover test

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello Frnds

Recently we had an audit of EDQM. One of its auditor asked me a question about HPLC but I was unanswered.
The question was HPLC carryover test at the time of HPLC calibration.
Does anyone have any idea about how to perform this test?

Thank you.
The manufacturer should be able to provide detailed instructions, but in general you run a high-level standard of a single compound and follow with a diluent blank. The area ratios give you the carryover.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 18 guests