Matrix Effects in LSMS Analysis of Plasma Samples
Posted: Sat Jun 11, 2005 10:01 pm
I have spent a lot of time trying to minimize the effects of matrix effects in the analysis of basic drugs in plasma. The major matrix interferences in samples analyzed by protein precipitation are phospholipids and lysophospholipids
I developed a method to employ one ion in MRM LCMS analyses that detects all the high-level lysophospolipids and phospholipids. I then used these to minimize the analysis time while still separating the basic drugs from these lipids. The basic useful parameters were flow, solvent composition, and temperature.
I also evaluated the effects on accuracy, precision, peak shape, retention time, column backpressure, etc when these lipids were not eluted from the column.
I was surprise to find that the signal for these lipids become "steady-state" in concentration. Also they began to elute with the analyte after many injections. After they became steady-state, and the analyses were not severely compromised.
I would be interested in others observations/comments:
see
ASMS Poster: "Simple Method to Monitor Lysophospholipids and Phospholipids During LC-MS Method Development via In-Source CID"
see
http://users.chartertn.net/slittle/default.htm
Top subject.
I developed a method to employ one ion in MRM LCMS analyses that detects all the high-level lysophospolipids and phospholipids. I then used these to minimize the analysis time while still separating the basic drugs from these lipids. The basic useful parameters were flow, solvent composition, and temperature.
I also evaluated the effects on accuracy, precision, peak shape, retention time, column backpressure, etc when these lipids were not eluted from the column.
I was surprise to find that the signal for these lipids become "steady-state" in concentration. Also they began to elute with the analyte after many injections. After they became steady-state, and the analyses were not severely compromised.
I would be interested in others observations/comments:
see
ASMS Poster: "Simple Method to Monitor Lysophospholipids and Phospholipids During LC-MS Method Development via In-Source CID"
see
http://users.chartertn.net/slittle/default.htm
Top subject.