Page 1 of 1
GC-MS to GC area conversion
Posted: Mon Sep 10, 2012 3:15 pm
by afmb
Im sorry if this query was already posted before. I tried searching through previous post but couldn't find anything regarding this. When i run samples in GC-MS, the area for chloroform that I obtained was 1,971,183,624. But when the samples were analyzed in GC, the area (pA*S) that i got was 341,144. Is there any way to convert the area that I obtained fr GC-MS to the area in GC or vice-versa?? What's the unit of area in GC-MS? TQVM for anyone who answers my question.
Re: GC-MS to GC area conversion
Posted: Tue Sep 11, 2012 2:00 am
by Don_Hilton
Response factors vary from detector to detector. The area for chloroform will vary in the mass spec depending on wether you sue the signal for one ion, another ion, or all ions (in the mass range acquired).
The best solution - build a calibration curve for each instrument.
There have been discussions of mass spec response relative to TIC on this site.
Posted: Thu Sep 13, 2012 9:23 am
by afmb
hey... thanx so much for the reply. Im doing the same thing as u suggested now. separate calibration curve for GC and GC-MS...