ELSD lowered response

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi.

I'm working on carbohydrates analysis on ZORBAX carbohydrate analysis column (4.6x250 mm 5 um). ELSD detection (70°C, 3.7 bar ). 75:25 ACN:water.

All was working well until we have to run another method which uses sulfuric acid as unique eluant (2mN) on an Aminex column. The signal response for the same concentration has suffered a 4-fold reduction. i.e Fructose at 200 ppm used to register a 6 mV signal, and now is registering less than 1.5 mV.

Sulfuric acid afected the detector? What should I do?

Thanks a lot
Clean the light scattering cell and avoid sulfuric acid in the future.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
You're right!
Thanks a lot.
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry