A quick question for all the HILIC experts out there.

When doing HILIC people often run gradients from around 5% water to around 50% water. I know if we go too much higher in water content we would cause the adsorbed water layer to be lost (dissolved into the mobile phase) but is 50% really the maximum?

I would think one could at least go to 60 or 70%?

Please let me know what your experience has been.

Much Thanks!!
Here are some rules of thumb:
1) To get good retention of your analytes, one usually needs about 70-75% organic solvent in the case of proteins, 85% ACN in the case of tryptic peptides, and 85-95% for small molecules that are only modestly polar. Exceptions include small molecules that are extremely polar, such as aminoglycoside antibiotics (e.g., kanamycin). To use higher levels of organic solvent than the concentration necessary for good binding risks precipitation of your analytes. For example, complex carbohydrates start to precipitate at 80% ACN, while many tryptic peptides aren't soluble at 90% ACN (but they are at 85%).
2) If you need 90+% to get retention of an analyte, then you usually need to run only to 50-60% organic solvent in a gradient to insure its elution. For a particularly heterogeneous mixture such as a tryptic digest or a crude extract of metabolites from a leaf, some of the components may not elute until you get down to ~ 40% ACN. In such cases it's prudent to run the gradient to 10-20% ACN.
3) With analytes that contain numerous polar binding sites, such as intact proteins or aminoglycoside antibiotics, run the gradient down to 5-10% organic solvent. It may also be necessary to run a simultaneously increasing salt gradient or to add a chaotrope such as 50 mM unbuffered formic acid or hexafluoro-2-propanol (chaotropes antagonize binding in HILIC, the opposite of their effect in RPC).

Hope this helps.
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
Mixed-mode HILIIC on Obelisc N column allows you to use extra 10-15% wtaer compare to traditional HILIC due to additional ion-exchange interaction (cation- and anion-exchange) for ionizable analyte. This allows you to avoid problem with solubility and mismatch between diluent and mobile phase (which is often the case for HILIC). It also increases loadability and retention for ionizable polar analytes. See our web site for details
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 3 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry