Dear sir
I think that I need more your help
i will provide you all the information about my sample and my method hoping that i may help you to provide a suggestion
my analyte are porphyrins in free acid form
what type of chromatography you're doing: reversed phase
the organic solvent: acetonitrile
the buffer: ammonium acetate
the buffer concentration: 0.1 M ( normallyI have to use 1 M of ammonium acetate but because i don't know about the effect on the column i tried to start with this concentration)
the *exact* column: Hypersil ODS C18
new column.
the temperature: 40 degree Celsius
what the sample is dissolved: in 0.1 M HCL
the sample concentration: I tried different concentration where i 've seen the tailing in all the chromatograme
what the injection volume is: 20uL
the tailing problem has been present consistently
the flow is 0.9ml/min
The pH is 5.16
I hope that those information will help
best
Maybe I´m a bit late but I also doing a work regarding porphyrins. I have no problem with tailing peakshapes BUT when i used HCl to dissolve the standard tube (I know that this one is suggested from the manufacturor) I got tailing and bad peakshapes. I use 6M formic acid instead and thar work awesome.
Best regards,
/Jonas