Maximum Water % When Doing HILIC
Posted: Fri Sep 07, 2012 2:46 am
by hajdaei
A quick question for all the HILIC experts out there.
When doing HILIC people often run gradients from around 5% water to around 50% water. I know if we go too much higher in water content we would cause the adsorbed water layer to be lost (dissolved into the mobile phase) but is 50% really the maximum?
I would think one could at least go to 60 or 70%?
Please let me know what your experience has been.
Much Thanks!!
Re: Maximum Water % When Doing HILIC
Posted: Fri Sep 07, 2012 1:45 pm
by Andy Alpert
Here are some rules of thumb:
1) To get good retention of your analytes, one usually needs about 70-75% organic solvent in the case of proteins, 85% ACN in the case of tryptic peptides, and 85-95% for small molecules that are only modestly polar. Exceptions include small molecules that are extremely polar, such as aminoglycoside antibiotics (e.g., kanamycin). To use higher levels of organic solvent than the concentration necessary for good binding risks precipitation of your analytes. For example, complex carbohydrates start to precipitate at 80% ACN, while many tryptic peptides aren't soluble at 90% ACN (but they are at 85%).
2) If you need 90+% to get retention of an analyte, then you usually need to run only to 50-60% organic solvent in a gradient to insure its elution. For a particularly heterogeneous mixture such as a tryptic digest or a crude extract of metabolites from a leaf, some of the components may not elute until you get down to ~ 40% ACN. In such cases it's prudent to run the gradient to 10-20% ACN.
3) With analytes that contain numerous polar binding sites, such as intact proteins or aminoglycoside antibiotics, run the gradient down to 5-10% organic solvent. It may also be necessary to run a simultaneously increasing salt gradient or to add a chaotrope such as 50 mM unbuffered formic acid or hexafluoro-2-propanol (chaotropes antagonize binding in HILIC, the opposite of their effect in RPC).
Hope this helps.
Re: Maximum Water % When Doing HILIC
Posted: Fri Sep 07, 2012 2:12 pm
by Vlad Orlovsky
Mixed-mode HILIIC on Obelisc N column allows you to use extra 10-15% wtaer compare to traditional HILIC due to additional ion-exchange interaction (cation- and anion-exchange) for ionizable analyte. This allows you to avoid problem with solubility and mismatch between diluent and mobile phase (which is often the case for HILIC). It also increases loadability and retention for ionizable polar analytes. See our web site for details