Query about c18 column

Posted: Wed Sep 05, 2012 6:20 am
by kshiprakapil
Dear all the members

can I use chloroform:methanol:diethyl amine in ratio of 90:10:0.2 respectively as mobile phase in c18 column for hplc analysis. I need some papers where chloroform was used as mobile phase with c18 column.

thanx and with regards.
Kshipra Kapil

Re: Query about c18 column

Posted: Wed Sep 05, 2012 9:34 am
by Gerhard Kratz
Before you can do your first injection, your C18 column is dead. No, you cannot use Chloroform in RP chromatography with 90% in the mobile phase.
Why you want to do that?

Re: Query about c18 column

Posted: Wed Sep 05, 2012 1:50 pm
by HPLCaddict
Before you can do your first injection, your C18 column is dead. No, you cannot use Chloroform in RP chromatography with 90% in the mobile phase.
Why you want to do that?
Ehm, really? I mean, I cannot think of a good reason why one wants to use Chloroform as eluent in RP-HPLC, but why should Chloroform kill the column?
As long as the 10% are not water :D it should work (in principal).

Re: Query about c18 column

Posted: Thu Sep 06, 2012 3:23 am
by tom jupille
Dichloromethane is occasionally used to flush junk from reversed-phase columns, so I wouldn't expect chloroform to damage anything. Miscibility with previous aquous solvents would be more of a concern.

Re: Query about c18 column

Posted: Thu Sep 06, 2012 7:28 am
by rhaefe
I don't have my copy in front of me but you might want to check out:
"The HPLC Solvent Guide" by Paul C. Sadek.
There is a chapter covering chlorinated alkanes and chlorinated benzenes.

rh

Re: Query about c18 column

Posted: Thu Sep 06, 2012 7:56 am
by Gerhard Kratz
In RP water or aquous buffer is in mobile phase with minimum 3-5% or more. And with regards to the imiscibility of Chloroform and Water this is the problem what will kill the column packing. Following the HPLC Solvent Guide of Paul Sadek to rinse the column with different solvents and at the end with Chloroform will take very long time and at the end it is cheaper to buy a new column.

Re: Query about c18 column

Posted: Thu Sep 06, 2012 1:37 pm
by HPLCaddict
In RP water or aquous buffer is in mobile phase with minimum 3-5% or more. And with regards to the imiscibility of Chloroform and Water this is the problem what will kill the column packing. Following the HPLC Solvent Guide of Paul Sadek to rinse the column with different solvents and at the end with Chloroform will take very long time and at the end it is cheaper to buy a new column.
Of course you have to carefully exclude water when using chloroform, but having water in the mobile phase is not a necessary prerequisite for RP-HPLC. I've seen methods with nonaqueous eluents (this is sometimes coined "NARP" - nonaqueous reversed phase). The methods I've seen used gradients of methanol and acetonitrile or THF, though, no chlorinated solvents.
So, using dichloromethane/methanol (there is no immiscibility between those two, if I recall correctly) on C18 might be a possibility when you're dealing with extremely hydrophobic analytes - I'd prefer using a less hydrophobic stationary phase with the usual RP-eluents :D .

Re: Query about c18 column

Posted: Thu Sep 06, 2012 1:45 pm
by HPLCaddict
Just to shoot out a quick-Google-example: Separation of carotenes by NARP using dichloromethane/ACN on C18:
http://onlinelibrary.wiley.com/doi/10.1 ... 5/abstract