Waters PDA 2996 Calibration

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi I am currently calibrating a Waters 2996 PDA detector, but it is failing for the 522nm peak the tolerance set is +- 1.5nm I am getting the peak at 524.1nm. I have replaced the Lamp but I am still getting the same results with the new Lamp. Any body any ideas about what could be causing the problem.

Thanks
The only thing I can think of that would cause that is a misaligned diffraction grating but I've never actually seen that happen and can't give any guidance as to how to fix it. It is probably not a do-it-yourself issue, unless there is a software fix.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
try to read the energy , if it low replace M1 and M2/3
If the problem is still exist, you must replace grating
The dirty flow cell window cause this phenomena. Wash or change the flow cell window.
4 posts Page 1 of 1

Who is online

In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry