Hello,

I am currently trying to determine a good sample concentration and split ratio for a GC method using a FID detector. Is there a peak height (milivolts)considered desirable with a GC-FID detector?

I know in HPLC a peak height around 1 AU is considered an appropriate height for accurate determinations. Is there a GC-FID version of this rule of thumb?

Thanks

James
GC differs from HPLC in the linearity of detection for the most common detector.

FID for most manufacturers is linear over a range of about six to seven orders of magnititude. HPLC UV detection is linear over a smaller range. Thus, the FID amplifier setting can and will be changed if conditions require it.

Measuring a peak which reaches the 1volt limit will require a split higher than a peak which is close to the detection limit which may require a splitless injection.

A peak having a height half the dynamic range of the amplifier should be measured accurately. Adjust your concentration and volume and split so this can be reached if conditions permit it.

I hope these meager comments help your understanding of the situation.

best wishes,

Rod
I know in HPLC a peak height around 1 AU is considered an appropriate height for accurate determinations.
1 AU is pretty darn large !!!! We usually stay under 100 mAU with today's better instrumentation.
I agree. Historically, getting near the 1 AU limit is out of linear range.

But one question at a time, right?

best wishes,

Rod
Thanks for your help Rod. I understand the situation a bit better now.

James
5 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry