Hi, could someone please clarify in which order would N2, O2, He, N2O, CH4 and CO2 appear on TCD when using He as carrier gas and a 5A molsieve column? Thanks, Hiz
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Hiz,
It is certain that you will not see He peak when using He as a carrier. I would be amazed if you did.
O2, N2, C1 would be the elution of these three peaks.
N2O and CO2 generally would not be detected as they are caught in the pores of the sieve and are eluted only with time and temperature as a flattened and distorted peak on most commercially available packings.
When I worked for Supelco I prepared MS5A packings that would allow the elution of CO2 as a symmetrical peak. I don't remember ever testing N2O on that packing, however.
best wishes,
Rod
It is certain that you will not see He peak when using He as a carrier. I would be amazed if you did.
O2, N2, C1 would be the elution of these three peaks.
N2O and CO2 generally would not be detected as they are caught in the pores of the sieve and are eluted only with time and temperature as a flattened and distorted peak on most commercially available packings.
When I worked for Supelco I prepared MS5A packings that would allow the elution of CO2 as a symmetrical peak. I don't remember ever testing N2O on that packing, however.
best wishes,
Rod
-
- Posts: 8
- Joined: Fri Oct 14, 2011 2:28 am
Dear Rod,
Thanks very much for the reply that explain why I wasn't seeing a N2O peak on TCD. Given we have to use molsieve for separating O2 and N2, what would you suggest to do for measuring N2O on TCD wwhile also using molseive? The reason why we have to use TCD for N2O is we expect high N2O concentrations and we cannot dilute these as these will be collected directly from sampling bottles.
Thanks
Hiz
Thanks very much for the reply that explain why I wasn't seeing a N2O peak on TCD. Given we have to use molsieve for separating O2 and N2, what would you suggest to do for measuring N2O on TCD wwhile also using molseive? The reason why we have to use TCD for N2O is we expect high N2O concentrations and we cannot dilute these as these will be collected directly from sampling bottles.
Thanks
Hiz
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
If I had to guess, a simple porapak Q would be as good as anything, but it might be a long one to separate CO2, and possibly subambient in temperature.
Good luck,
Rod
Gas Chromatograph Separation of Oxides of Nitrogen.
J. M. Trowell
Anal. Chem., 1965, 37 (9), pp 1152–1154
DOI: 10.1021/ac60228a021
Publication Date: August 1965
Gas chromatographic method for measuring nitrogen dioxide and peroxyacetyl nitrate in air without compressed gas cylinders
Mark R. Burkhardt , Nyengenya I. Maniga , Donald H. Stedman , Richard J. Paur
Anal. Chem., 1988, 60 (8), pp 816–819
DOI: 10.1021/ac00159a017
Publication Date: April 1988
Good luck,
Rod
Gas Chromatograph Separation of Oxides of Nitrogen.
J. M. Trowell
Anal. Chem., 1965, 37 (9), pp 1152–1154
DOI: 10.1021/ac60228a021
Publication Date: August 1965
Gas chromatographic method for measuring nitrogen dioxide and peroxyacetyl nitrate in air without compressed gas cylinders
Mark R. Burkhardt , Nyengenya I. Maniga , Donald H. Stedman , Richard J. Paur
Anal. Chem., 1988, 60 (8), pp 816–819
DOI: 10.1021/ac00159a017
Publication Date: April 1988
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
I have an article (without attribution, sorry) by Zechmeister-Boltenstern which discusses N2O by EC and TC. I may have found it via web search. Uses Porapack QS column (along with sieve in multi-valve configuration.)
Best regards,
AICMM
Best regards,
AICMM
5 posts
Page 1 of 1
Who is online
In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 19 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 19 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science