PRP-X110S Suppression
Posted: Thu Aug 30, 2012 1:08 am
by vockhead
I was wondering if anyone had any experience using a membrane suppressor with the PRP-X110S column? I'm using a 4 mM parahydroxybenzoic acid, 5 mM NaOH, and 0.1 mM sodium thiocyanate eluent in 2.5% MeOH, but am trying to figure out the appropriate regenerant.
I'm using a Dionex AMMS 300 suppressor, which states that the dynamic concentration through the suppressor for eluents such as NaOH and HCO3/CO3, is to multiply the normality of the mobile phase by 7 to figure out the dynamic normality of your regenerant.
The only thing is, those eluents are all bases, while the PRP-X110 column has an acid within the eluent, which I'm guessing can't be suppressed. So would the normality of my eluent by 5 mN NaOH and the 0.1 mN thiocyanate and ignore the parahydroxybenzoic acid as it is an acid already?
Any help would be greatly appreciated,
Vick
Re: PRP-X110S Suppression
Posted: Fri Aug 31, 2012 8:59 am
by Markus Laeubli, Metrohm
Dear vockhead
In your eluent the parahydroxibenzoic acid is titrated by NaOH. Your eluent composition is (inorganic part):
4 mmol/L sodium parahydroxibenzoate
1 mmol/L sodium hydroxide
0.1 mmol/L sodium thiocyanate
This makes up to 5.1 mmol/L sodium which has to be suppressed. The anions are only of interest as after suppression the dissozation of the respective conjugated acids add to the background conductivity.
Posted: Tue Sep 04, 2012 2:42 am
by vockhead
Hello Markus,
You were dead on, thanks for your help.