I squeezed an air bag to inject a gas STD into a sample loop system. The gas STD including benzene, toluene, ethylbenzene and p-, m-, o-xylenes (only 10 ppm for each compound), balanced with Nitrogen. There is only benzene peak showed up. Toluene sometimes showed up with weak intensity. All others never showed up. If I manually inject the STD from injection port by using a 50 micro-L syringe, I can see all compounds. My question is, where are those missing compounds when using sample loop system?
The loop system is a little complicated:
1. When filling the loop by squeezing an air bag, three loops are supposed to be filled: Loop 1 (1 mL) -> Loop 2 (0.5 mL) -> Loop 3 (0.5 mL) -> vent
2. When start the GC program, a 10-way and 3 6-way valves are turned and gases in the loop system are carried into the GC columns
Loop 1 -> Hayesap series columns -> TCD
Loop 2 -> Rt-Alumina column -> FID 1 (split ratio = 50)
Loop 3 -> Rtx-1 column -> FID 2 (split ratio = 39)
Aromatics are expected to be shown in FID 2
The loop system is in a heating box with 160 (also tried 90) degree C. So, it is impossible that these aromatics are condensed. Oven temperature = 90 degree C and is programmed to increase to 200 degree C.
Benzene peak looks good and reproducible, but not for the others.
GC: Shimadzu 2014
Carrier gas: Helium
Injection port T = 150 degree C
Mode: constant pressure (due to high pressure, I only can use constant pressure... also tried constant flow but didn't help)