Method Accuracy
Posted: Thu Jun 09, 2005 10:51 am
We have developed and validated a method with an accuracy acceptance criteria of 80 to 110% recovery (for actives in animal feed). The recovery value was consistent across several different concentration and was around 85%. As we have shown in the short term that this recovery is repeatable we are planning to introduce a correction factor to bring this up to 100%. We would very long term accuracy by running an accurately prepared sample with each analysis.
Has anyone performed anything similar or have any comments?
Thanks
Has anyone performed anything similar or have any comments?
Thanks