Compound is VERY insoluble in water, soluble in most organics though. Acetic Acid is one of the necessary components to evaluate. Trying most solvents at 100ppm for validation, benzene at 1ppm. Acetic Acid is the only one giving fits. Given the water insolubility of the compound (it doesn't work with most HPLC options), our other options for evaluation are small, we don't routinely do NMR.
Rtx-stabilwax, 30x0.53x1. DMF diluent, direct 180*C splitless injection, 200*C FID detection, tapered glass wool liner, 1uL injection. Went with External Standard because the internal standard seemed to be an issue (maybe jumped off that bandwagon too soon though). The standards look fabulous, all peaks <2% RSD over 6 injections.
Move on to samples and accuracy/recovery: All compounds are recovered at 100 +/- 10% EXCEPT Acetic Acid. Now usually I would assume consumption but it's actually coming out at 200% recovery. There is none in the blank sample injection.
Any ideas as to why it's showing up so high? I was going to try validating at a higher ppm limit, say 500 ppm or so and see if it helps, we have up to the 5000 ICH limit but you know consumers, they want to see none so we need it validated lower if at all possible. Or using an acidified diluent? Injector/detector temps?