I wonder if this variation should be a concern, or I can simply ignore it.
Thank you.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
I'm currently using an Agilent 1100 HPLC eqiupped with a binary pump, autosampler, and DAD. In order to assess reproducibility of reults, I injected successively content of a vial for 6 times, and it resulted to a mean peak area of 508.52 mAu, range from 489.3 to 528.4, standard deviation 14.86, and CV% of 2.92. I also see an increasing trend in this data set.
I wonder if this variation should be a concern, or I can simply ignore it.
Thank you.
I have exact system. Our repeat injections meet system suitability requirements of < 2%RSD essentially 100% of the time, even with gradient elution. So something is wrong. I agree to start with the separate vials.I'm currently using an Agilent 1100 HPLC eqiupped with a binary pump, autosampler, and DAD.
I wonder if this variation should be a concern, or I can simply ignore it.
Thank you.