I am Maharshi from Gujarat.
I am using JASCO PU1550 HPLC instrument with manual injector.
I've problem with the retention time in analysis of pharmacopoeial drugs.
One of my API is Adrenaline Bitartrate. I am using C18 RP-II 100X4.6mm, 3 micron column as per the pharmacopoeial requirement. Program is binary gradient and also as per the requirement and already setted in the instrument.
pH of mobile phase A is 2.2 and of mobile phase B is 2.4 as per the requirement.
One of our HPLC engineer advised us to reduce the length of tubings from injector to column and column to detector.
Before changing the length of tubing, I was getting the proper retention time. But after changing it to shorter length and small diameter, the retention is changed. Previously it was about 4 as per monograph but now I am getting about 1.4.
Not only this product but I am getting the same problem with the other products also.
Then I changed the tubings and again set the old one but still its not solved.
Column is also a new one. I've also verified it on old column but no difference I've got.
Now I am unable to understand what could be the problem.
Anyone please help me.