Formic Acid on an Agilent HP-5 Column?
Discussions about GC and other "gas phase" separation techniques.
18 posts
Page 1 of 2
One component of the samples I need to analyze is Formic acid. Can I run small amounts (~3mg/ml) of formic acid through this column without activating it?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
It won't do the column any good.
Film thickness has a big bearing here.
Just how do you intend to detect formic acid with your column?
Rod
Film thickness has a big bearing here.
Just how do you intend to detect formic acid with your column?
Rod
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
How do you intend to convert all the formic acid to a free acid form? salts will NOT work for you.
Rod
Rod
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I've successfully used HPLC/IC with an organic acid column and conductivity detector for formic acid assay.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
The method used by CPG is MUCH PREFERRED over GC.
Unless you know what you are doing, GC is NOT recommended.
Rod
Unless you know what you are doing, GC is NOT recommended.
Rod
-
- Posts: 25
- Joined: Tue Jul 03, 2012 2:33 pm
Ok, thanks for the advice. I figured I would just detect formic acid via FID, but it sounds like this is a moot point. The film thickness is 0.25 um, in case that makes a difference.
Any new advice with this information?
Any new advice with this information?
-
- Posts: 37
- Joined: Thu Jul 19, 2012 6:21 pm
Restek and others have some chromatograms analyzng derivitized formic acid by GC/FID or GC/MS.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
You will NOT be able to detect formic acid by FID.
Sorry,
Rod
Sorry,
Rod
-
- Posts: 25
- Joined: Tue Jul 03, 2012 2:33 pm
Ok, thanks for all your helpful advice. I think I will just stick with a liquid/liquid extraction and titration for now.
-
- Posts: 37
- Joined: Thu Jul 19, 2012 6:21 pm
Not with this column, but Phenomenex has a method for detecting the methyl esther here: http://www.phenomenex.com/Application/D ... ias=formic acid&returnURL=/Application/SearchYou will NOT be able to detect formic acid by FID.
Sorry,
Rod
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Correct, one has to derivatize the formic acid to see it by FID. The free acid does NOT detect with a FID.
best wishes,
Rod
best wishes,
Rod
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
What is the issue there, Rod? Little or no FID response? Or the free acid cannot make it through the inlet/column/detector?
Thanks.
Thanks.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Well, there is almost no response with an FID.
But getting the acid there with a non-free acid phase is quite a difficult matter.
HID or TCD will work for detectors. But on column injection, followed by a free fatty acid phase can elute the free acid.
But not on a 5% phenyl methyl silicone phase.
best wishes,
Rod
But getting the acid there with a non-free acid phase is quite a difficult matter.
HID or TCD will work for detectors. But on column injection, followed by a free fatty acid phase can elute the free acid.
But not on a 5% phenyl methyl silicone phase.
best wishes,
Rod
-
- Posts: 25
- Joined: Tue Jul 03, 2012 2:33 pm
Why is it that there's no response from an FID? Is it because the formic acid won't Ionize?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
There are still issues with exactly HOW an FID works. I have read variations on the mechanics of the detector and I don't know if anyone has nailed it.
Be that as it may, certain gases in a hydrogen flame just don't give much or any of a response. I prefer to think that formic acid is methane that is almost completed oxidized already when in its formic acid form. Your proposal that it does not ionize is as good as any I have heard. Or if it does ionize it does not conduct electricity well in gaseous state between two electrodes.
Life is chemistry and chemistry can be a pain. Everything else is commentary.
best wishes,
Rod
Be that as it may, certain gases in a hydrogen flame just don't give much or any of a response. I prefer to think that formic acid is methane that is almost completed oxidized already when in its formic acid form. Your proposal that it does not ionize is as good as any I have heard. Or if it does ionize it does not conduct electricity well in gaseous state between two electrodes.
Life is chemistry and chemistry can be a pain. Everything else is commentary.
best wishes,
Rod
18 posts
Page 1 of 2
Who is online
In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 15 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science