CP 3800 VARIAN GC DISCONNECTION
Posted: Wed Aug 15, 2012 11:41 am
by kuziwa
Hello.
I'm having a trouble with my GC; it cannot do continuous (multiple) analysis.
Once a single chromatographic run is completed, the GC stabilize and go to "not ready"
status, while from the workstation (PC) screen the module (GC) says it is "ready" and disconnect
as soon as you try to run another analysis. But if you turn the GC off and on again, it works fine for that single analysis.
So you have to keep turning it OFF & ON for every single run.
Please advice.
Kuziwa. A.K
Re: CP 3800 VARIAN GC DISCONNECTION
Posted: Tue Sep 11, 2012 3:00 pm
by fendersim
I would try another PC if you can, or failing that it might the communication board on the GC itself.
Posted: Tue Sep 11, 2012 4:47 pm
by WillyOne
I suggest you to check the instrument configuration in workstation. The 3800 icon should be placed into the rectangle.
Posted: Wed Sep 12, 2012 5:37 am
by kuziwa
Thank guys, let me try that.
Posted: Wed Sep 12, 2012 8:14 am
by Andy F
A different, but similar situation we've had is when the Not Ready, Ready and Run light are all illuminated simultaneously, very pretty, but not very useful. The solution we've found is to get the GC up before booting the Workstation PC up. It may help with your problem.
Posted: Wed Sep 12, 2012 9:27 am
by kuziwa
Thanks Andy, but it is still disconnecting after the run.
Any other idea, please?
Posted: Thu Jul 30, 2020 4:44 am
by aarthimohan
Hi everyone,
We have a Varian 3800 GC. When turned ON, the fans turn ON. However there is no communication with the software and the LCD display shows a broad horizontal line on it. Nothing else comes up on the display. It's looks like it is not a power supply issue as the 5V LED light on the main board comes up.
We are thinking both the LCD display unit and the COM board need replacement. Any thoughts you might have are appreciated. Really on a time crunch here. Please do help!
P.S.: Trying to wrap up an important project and also get my degree! The instrument is the heart! If you are able to provide some assistance over a call even better.
Posted: Wed Aug 05, 2020 11:31 am
by HbJ
Please check for any CR2032 cells, especially on the mainboard.
They might be empty, corrupting the NVRAM contents.
Exchange those and restart.