why is empower only injecting one injection from sample set?
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
11 posts
Page 1 of 1
im running a sample set and it is only injecting the first injection . it doesnt continue and inject the rest of the sample set...why is that?
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
any messages in your message center that may help?
-
- Posts: 2174
- Joined: Tue Aug 17, 2004 7:59 pm
Did you specify a run time in the sample set method?
Thanks,
DR

DR

-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
well i have a gradient set up that stops at 15mins but the injection just keeps running....any ideas why?
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
I still havent solved this one guys...please help
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
what runtime did you set in your sample set?
Do you use events in your instrument method?
Do you use events in your instrument method?
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
my run time is 15mins in the sample set.....i havent set any events...should i?
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
ive talked with the technical support and they seem a bit clueless....
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
No you shouldn't, but just in case you had set some....my run time is 15mins in the sample set.....i havent set any events...should i?
I think we need some more information:
Does your sample set stops after your first injection, or just keeps running without recording and going to the next injection?
Do you recieve any messages, in Empower's message center, or the windows event viewer?
What kind of instruments are we talking about: Alliance, GC, UPLC,....
Is this connected to a PC or a LACE?
Stand alone or Enterprise configuration?
Ace
-
- Posts: 2174
- Joined: Tue Aug 17, 2004 7:59 pm
Add an event at 0.01min for the autosampler, turning switch 1 on. Move your start cable from the inject start over to switch 1. If the autosampler is not sending the start signal beyond the first injection of the sequence, this may work...
We had something similar going on w/ one of our 2695s and this worked. After a couple of days, we had a service engineer appt. setup and we put it back. It worked again once the engineer showed up - and has been working since. That's a good thing, because that set of contacts is attached to the main system board (expensive to replace).
We had something similar going on w/ one of our 2695s and this worked. After a couple of days, we had a service engineer appt. setup and we put it back. It worked again once the engineer showed up - and has been working since. That's a good thing, because that set of contacts is attached to the main system board (expensive to replace).
Thanks,
DR

DR

-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
ok got its solved i believe.waters engineer installed a cable between back pda and sample manager.He said that normally you dont need this extra cable for aquity but in this case it did.cable is inserted into pda inject start jack. Thanks everybody for there help
11 posts
Page 1 of 1
Who is online
In total there are 48 users online :: 1 registered, 0 hidden and 47 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 47 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 47 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science