Problem in Empower (2707 A.S + 1525 Pump + 2998 Detector)

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

6 posts Page 1 of 1
Dears all

I have some problems with empower with waters auto sampler 2707:

During the run of sequence, the empower software is suddenly stop and abort injections, actually many times do this , so i do the following:

1-I check up cables and every thing is okay

2-I make "ping" to ip address for 1525 pump and detector and 2707 and found for 1525 pump:
C:\ping 127.0.0.1

Pinging 127.0.0.1 with 32 bytes of data:
Request timed out.
Request timed out.
Request timed out.
Request timed out.

Ping statistics for 127.0.0.1:
Packets: Sent = 4, Received = 0, Lost = 4 (100% loss),

so I think the problem may be either in firmware or communication


any suggestions???????????????????????
127.0.0.1 isn't the IP adress of your pump, its your local host.

any messages in your message center?


Ace
IP addresses of components are best found in node properties, DHCP config. screen.
They should all be 192.168.0._

Ping those addresses.
Thanks,
DR
Image
yeah, you right DR and ACETO_81 .. but generally, the problem of 2707 auto-sampler which occurring many times when doing run samples sequence and suddenly the run is interrupted or stopped .. which I must make restart for all HPLC and PC and start my sequence again and sometimes the problem repeated again many times.


note: I think this problem are mainly occur to many labs having 2707 A.S waters



any suggestions?????????
Any anti Virus/malware software on the computer? Does it happen at approximately the same time each day? Have you checked the PC's system and application event logs?
Thanks,
DR
Image
1- there is anti virus software and updated to 2012
2- no, this happen at different times, approximately each weak may happen twice and sometimes it happen daily.
3- I checked PC's Systems ,and systems PC are good
4- Also, Firmware version I checked it up, and its OK

note : this problem is mainly occur at different labs at different locations , so I think may be this problem due to communication or defective in firmware .. Actually, am don't know
6 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry