to BMU_VMW: I've probably made microliter injections of DCM into an inlet that hot and never given it a second thought.
An inlet at 300 and 50 microliters - that is a lot of temperature and a lot of solvent. With the volume, I assume this is not an injection into the normal split/splitless inlet unless you have a precolumn in a GC oven cool enough to condense the DCM and wide enough to handle the rapid gas transfer from the inlet.
You do not indicate wether you are talking about injection for a packed column or a capillary column. So, I 'll keep rambling on thinking in terms of cappilary.
Dow, on their website notes that DCM decomposes above 120 deg C in the presence of metal surfaces. With a 1 microliter injection made with a fast injection (typical of many autosamplers) the presence of the needle would not be a problem - it is not going to heat up that fast. To inject 50 microliters, a fast injection into a 300 degree inlet is going to generat a lot of vapor - and as fast as it is injected, so to avoid streaming vapors into the inlet you would, most likely have to do a slow injection - which would allow the needle to heat. This may or may not be sufficient to be a bit of a problem.
However I'd like to step back and ask why 300 degrees? I do have a few types of samples that have requried an inlet this hot, but some types of molecules are destroyed at this temperature. I'd rather to a cool injection in a PTV inlet, venting the solvent, and then ramp the temperature up, transfering the analytes to the column. This is easier on some of the inlet components and on the more labile analytes in the mixture.
to sonicboom12345: Methanol does condense, but in scattered droplets down a methyl slicone column, not a nice band. So, an injection above the boiling point of methanol works much better. It's been a while since I've set up anything using methanol as a solvent into the GC. I think I used something like 85 degrees. It may have been a bit cooler. DCM, I have injected as low as about 40 degrees.
When you do try for solvent focusing, keep the temperature below the boiling point of the solvent for a minute or so, allowing the sovent to gently evaporate and focus the band. Then, ramp up to what you need to elute your analytes.
Having said that, the longer the analytes stay on the column after the solvent elutes, the less solvent effect you have. So in the analysis of high boiling compunds you spend a lot of time cooling, focusing, and ramping up for little or no effect. In that case, inject into a hot oven - get the job done, and move on.