ghost peak
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
hi i am getting a ghost peak after the peak of interest have tried new prep brand new column and cleaned up the hplc system with nitric acid the peak doesnt change size or retention time at all what should i do i have been performing the test in years with no problem
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Isocratic or gradient?
And does the peak show up in blanks or in samples only?
And does the peak show up in blanks or in samples only?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3
- Joined: Thu Aug 09, 2012 5:46 pm
It is an isocratic method and it only shows up in the sample injections the column is a polysep GFC 1000 and mobile phase 10% methanol in water.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
If you're not already doing so, try dissolving your sample in the mobile phase and see if the peak goes away. If it persists, then it's possibly real, or due to some sample degradation. From the column, I surmise that it's a size exclusion separation, which would mean that the ghost is a smaller molecule than the main peak (or else that there is some secondary interaction going on).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 52
- Joined: Tue May 05, 2009 5:24 am
Is your blank taken through all the same sample preparation steps as the sample?
e.g. pipetting, filtering, centrifuging, etc? Changes in the quality of your sample prep consumables could be the cause
e.g. pipetting, filtering, centrifuging, etc? Changes in the quality of your sample prep consumables could be the cause
5 posts
Page 1 of 1
Who is online
In total there are 341 users online :: 1 registered, 0 hidden and 340 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 340 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 340 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science