Dear all,

I inherited a Shimadzu CTO-6A column oven. This unit was controlled with a SCL-6 controller. Unfortunately I don`t have this controller or a manual.

My problem is that the unit won`t heat up. I strongly believe this is due to the missing controller. I know there is a "blind plug" for uncontrolled systems, but I don`t have this item.

Perhaps anybody is still working with this old dinosaur and is able to have a look at this plug. It would be nice to get the pin out for this plug so I could make my own.

Thank you in advance.

Best regards

Stephan
Just short pins 1 and 2 (on the blue remote connector of the unit, counting from left).
Earth the rightest pin too (that's how Shimadzu's blind plug looks like).
2 posts Page 1 of 1

Who is online

In total there are 62 users online :: 2 registered, 0 hidden and 60 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 60 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry