Noticing a ripple in my chromatograms and just wondering what might cause this? Sample solvent is heptane and using FID, DB5 column, split 20:1 and He carrier. The ripple is noticable when the temp starts to increase from 100 to 300 (10C/min)

TIA
A couple of things:
1) if you could post a picture, it may help
2) are all gas pressures from the cylinders at least 20 lb above the pressures set on in the GC? (Common cause of a ripple with a time constant of a few minutes is regulators attempting to control with too low a pressure drop across them.)
Yes, a picture would help. Do you have a high purity regulator with stainless steel diaphrams? What grade of He are you using and how old are your purifiers? Make sure your regulator cylinder valves are fully open. If they are partially open the flow can fluctuate and sometimes stop completely.
3 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry