Headspace FID negative peaks

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
Dears,

I have a problem with reproducible negative peaks during analysis of residua solvents.
The system consists of Turbomatrix HS40 and a Shimadzu GC2014(FID).

The peaks are not present when injection with gas-tight syringe is performed.
Please advise what the cause might be.

You can see a chromatogram of blank water here:
Image
If you could give a bit more detail on what you are doing, it might help.

The only way to generate negative peaks in FID is to have a background level of something in the carrier stream that generates signal. And that background being displaced by something that gives no signal (like CO2, N2, Ar...).

So if you could tell what you are using for carrier gas and gas associated with you head space sampler. And then instrumental conditions for the sampler and for the GC, including what column you are using would help as well. If you could tell us what kinds of samples you are analyzing, that may give a clue as well.
I hope the details would help:

The system is used for various analyses of residual solvents in pharmaceuticals.
Solvent: Water
Carrier gas & Heads pace gas: Nitrogen
Column Elite-624 60m

HS...........
Thermostatting time: 20 min
pressurization time: 1 min
withdraw time 0.2 min
Inject volume 1 mL
Needle tepm. 85 degrees
Transfer line tepm. 115 degrees

GC
Injector temp 180 degrees
Detector temp 240 degrees
Flow rate ~ 3mL/min
Program: 40-200 degrees in 6 minutes

The whole system was conditioned for 8 hr (capped detector).

Thanx
On a 60 m column with nitrogen as carrier at the optimum flow rate the dead time should be about 6.5 min, which would mean that the whole chromatogram shows the period before the first peak elutes. You have a flow rate of 3 ml/min, what is the internal diameter of the column, and what is the inlet pressure ?

Peter
Peter Apps
I may be wrong with the flow rate.

I'll double check it
I may be wrong with the flow rate.

I'll double check it
A very good idea :x

If you are wrong about the basic parameters of the analysis, how right is the advice you get going to be ?

Peter
Peter Apps
6 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry