hi,i really need help
all i know is how to run the sample and get results.we are testing butan (propan,i-butan,n-butan,i-pentan,n-pentan).we use agilent 6890,chemstation gc,injection manuelly with sample cylinder.
what i have to know is INTEGRATION and CALIBRATION. we have standard sample.
do i always have to do integration before calibration?
how many standard samples do i have to run?
How do i get average?
can anybody explain but step by step :) integration and calibration

Re: integration,calibration...help :)

Posted: Mon Aug 06, 2012 9:47 pm
by chromatographer1

what i have to know is INTEGRATION and CALIBRATION. we have standard sample.


do i always have to do integration before calibration?

YES

how many standard samples do i have to run?

It depends on the regulations under which you are working. One is a minimum. Six would be too many????

How do i get average?

Add the results from the runs for each peak and divide by the number of runs.


can anybody explain but step by step :) integration and calibration


Integration is measuring each peak for area. Calculating the area of each component at a certain concentration gives you a calibrated response for that concentration.

Comparing areas of an unknown concentration sample with the standard sample allows you to calculate the concentration of the unknown sample.

I hope this helps.

Rod