I maintain an HP 6890/5971A GC/MS system where I study and am stumped concerning a recent problem. Recently the MS stopped collecting data and the log displayed the following messages (paraphrased):
Diffusion pump is too hot
Diffusion pump is too cold
System is in VENT state
Run Error: 1101,"(RUN:RDY ON;): Mass spec fault during acquisition : 417"
The detector manual and error code search didn't provide much insight. I noticed the MS was making an unusually loud noise and that the pressure reading on the ionization gauge controller was ~10^-3 torr, about 100 times higher than normal. The diffusion pump fan is functioning and the noise seemed to come from inside the rear of the MS. Upon checking the vacuum status, the pressure was normal (50 mtorr). I'm assuming that the pressure value in the vacuum status window is the foreline pressure, while that from the gauge controller box is the source pressure? Anyway, when I tried to vent the system the program seemed to be unable to complete the vent cycle. I'd really appreciate any feedback on possible causes/remedies for the problem. Thanks.
-Pat N.