- Posts: 15
- Joined: Thu Feb 24, 2011 2:49 am
We are having problems with our 5973 MS which is connected to a 6890 GC. Initially we were having trouble with an increase in EMvolts and an air leak. We found the airleak, re-cleaned the source and did a big clean/maintenance on the machine. Now we are having trouble with the auto-tune. The tune runsfine until it gets to the ion focus lens/repeller/entrance lens section and doesn't give any parameters and therefore increases the EM-volts at the end of the tune trying to find the PTFA peaks. (We abort the tune when we see this so as to not ruin our filaments or EM horn)
When we go into the manual tune we put parameters into the manual tune which we got from an auto-tune which previously looked good. When we try ramping the entrance lens/repeller/ionfocus lens we get straight lines:
However, the scan and profile are fine:
All of the cords connecting the source are correctly connected.
Any help would be greatly appreciated