99% loss of sensitivity in GC/MS!!!
Posted: Thu Aug 02, 2012 4:54 pm
Hi all I have a problem I can't see anything except major solvent peaks on my GC/MS
A bit of history: I started a new position and found a GC/MS Agilent 6890-5975 in terrbile shape due to the gas line plumbing. It had worn compression fittings cobled together with ptfe tape and a bad mix of brass ferrules on steel lines and mix and match cu and steel tubing. It had bad leak problems. I redid all the lines and got it very nicely air tight. It worked well for a while.
Last Friday I did some PM, I cleaned the source, replaced the liner, septa, gold deal, clipped the column on both ends, replaced the PV trap. The bottom filament is in rough shape but all in it looked fine when I got it back together. It passed the atune em volts are getting a bit high 2000-2200 but the peak splitting on the 69 and 219 dissappeared after I cleaned the source.
Anyways now the main issue. I ran our 3-mcpd calibration and found the peaks had poor resolution but the correct retention times and good linearity. However the peak areas were 1% of what they were several months ago. I reclipped the column (thought It may be positioned wrong or cut badly), checeked the autosampler, and reran the atune and it was fine. So I reran the calibration and the same result.
I didn't have a performance check standard handy so I made my own. I grabbed some reference standards
(2-methyl butyrate, R-Limonene, 2,3 heptane dione, 2-furyl methly ketone, and diethyl-l-tartarate, and made a 4 ppm / ug/ml solution in n-hexane and injected 1ul splitless. I only saw the solvent peak and nothing else.
I figure the column is probably toast from the leaks but is that enough to cause this. Please give me some suggestions. Thanks
A bit of history: I started a new position and found a GC/MS Agilent 6890-5975 in terrbile shape due to the gas line plumbing. It had worn compression fittings cobled together with ptfe tape and a bad mix of brass ferrules on steel lines and mix and match cu and steel tubing. It had bad leak problems. I redid all the lines and got it very nicely air tight. It worked well for a while.
Last Friday I did some PM, I cleaned the source, replaced the liner, septa, gold deal, clipped the column on both ends, replaced the PV trap. The bottom filament is in rough shape but all in it looked fine when I got it back together. It passed the atune em volts are getting a bit high 2000-2200 but the peak splitting on the 69 and 219 dissappeared after I cleaned the source.
Anyways now the main issue. I ran our 3-mcpd calibration and found the peaks had poor resolution but the correct retention times and good linearity. However the peak areas were 1% of what they were several months ago. I reclipped the column (thought It may be positioned wrong or cut badly), checeked the autosampler, and reran the atune and it was fine. So I reran the calibration and the same result.
I didn't have a performance check standard handy so I made my own. I grabbed some reference standards
(2-methyl butyrate, R-Limonene, 2,3 heptane dione, 2-furyl methly ketone, and diethyl-l-tartarate, and made a 4 ppm / ug/ml solution in n-hexane and injected 1ul splitless. I only saw the solvent peak and nothing else.
I figure the column is probably toast from the leaks but is that enough to cause this. Please give me some suggestions. Thanks