Is it possible to run gc and lc on one mass spectrometer after changing the source,
Thanks for your answers
Is it possible to run gc and lc on one mass spectrometer after changing the source,
Thanks for your answers
The most common ion source for GC-MS is electron impact. Sample is introduced in the gas phase coming off the capillary column.
The most common ion source for LC-MS is electrospray. Sample is introduced as a liquid coming off the LC column.

What exactly are you proposing?
Is it possible to run gc and lc on one mass spectrometer after changing the source,
Thanks for your answers
Yes it is.. I have such a setup... But for GC you have to an APCI ( OR water's version called APGC) source, NOT EI.

- Karen
Thank you Karen I will check
Varian had the 1200L MS, which with a source change could be either an APCI or ESI-LCMS or an EI or CI-GCMS. My lab uses ours as an LC-ESI-MS/MS, but if we had the hardware to interface a GC, we could also run GC-MS or GC-MS/MS on the same box. It's not the most sensitive instrument out there, but it's definitely good for routine work (ppm to mid ppb-level detection of compounds in a soil or water extract, for example. which depending on the sample concentration factor could result in ppb to ppt level reporting limits), and most definitely an academic environment if all the equipment is present to switch between LC and GC-MS.
Time flies like an arrow. Fruit flies like a banana.
5 posts Page 1 of 1

Who is online

In total there are 54 users online :: 0 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 54 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry