- Posts: 41
- Joined: Thu Jan 26, 2012 2:43 pm
Just can´t find exact description why is saponification needed for example for the determination of vitamin A and E with HPLC analysis. Can anyone explain it or point to a source where I can get the info?
Advertisement
Discussions about sample preparation: extraction, cleanup, derivatization, etc.
The method is for foodstuffs in general and same sample preparation is for vitamines D2 and D3, alpha-,beta-,gamma- and delta-tocopherols and for vitamin A. So it is meant for example to transfer Retinyl palmitate to retinol? Also does the the degradation of fatty matrix serve any purpose? Is it bad for RP-column?looking around on the internet, I see some methods that include saponification and others that do not. This appears to be tied to the objective of the method and the matrix.
These vitamins may be present as esters. To obtain a total content, it may be approporatie to saponify the esters to have all of the vitamin present as the free alcohol.
Saponification can be used to degrade a fatty matrix.
So to answer the question regarding the the analysis you are using, it might be appropriate to know what matrix and if you have the possibility of esterified vitamins.
It would appear that this is the case. However, the best way to find out the intent of the method is to find out what the intent of the method was when developed or selected. A properly documented method should state analytes of interest and matricies for which it is intended. If this is a method that was handed to you, you may have the procedural steps that were extracted from a larger document or set of documents. The answer to this means tracking back the source of your method.So it is meant for example to transfer Retinyl palmitate to retinol?
Yes. If the saponification was intended to degrade fatty matrix, it would be intended to improve the analyis in some way. This could be to remove interfering peaks in the chromatogram. It could be to improve peak shape. It could be to... Well, without details about the method, I can make a lot of guesses.Also does the the degradation of fatty matrix serve any purpose?
This may be better answered by a person who does LC, but I can say that this depends on what mixture goes on the column. Again details would be helpful.Is it bad for RP-column?
Most triclycerides are not too soluble in solvents like methanol or ACN. That's why saponification of such samples is commonly done.[This may be better answered by a person who does LC, but I can say that this depends on what mixture goes on the column. Again details would be helpful.Is it bad for RP-column?
If you can put the specific method steps here, someone running a similar method may be able to address specifics.
I found them all but only managed to work through vitamin D. The method was validated for drinks and foods in general(does not mention any specific matrixes used(seems wrong to me, but I don´t have alot of experience so it could be ok)). The method was validated for sum of vitamin D(D2+D3). I also found the document which the method was based on and it was european standard EVS-EN 12821It would appear that this is the case. However, the best way to find out the intent of the method is to find out what the intent of the method was when developed or selected. A properly documented method should state analytes of interest and matricies for which it is intended. If this is a method that was handed to you, you may have the procedural steps that were extracted from a larger document or set of documents. The answer to this means tracking back the source of your method.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.