Problem in Empower (2707 A.S + 1525 Pump + 2998 Detector)
Posted: Sun Jul 29, 2012 4:51 pm
by sherif4chem
Dears all
I have some problems with empower with waters auto sampler 2707:
During the run of sequence, the empower software is suddenly stop and abort injections, actually many times do this , so i do the following:
1-I check up cables and every thing is okay
2-I make "ping" to ip address for 1525 pump and detector and 2707 and found for 1525 pump:
C:\ping 127.0.0.1
Pinging 127.0.0.1 with 32 bytes of data:
Request timed out.
Request timed out.
Request timed out.
Request timed out.
Ping statistics for 127.0.0.1:
Packets: Sent = 4, Received = 0, Lost = 4 (100% loss),
so I think the problem may be either in firmware or communication
any suggestions???????????????????????
Re: Problem in Empower (2707 A.S + 1525 Pump + 2998 Detector
Posted: Mon Aug 13, 2012 2:05 pm
by aceto_81
127.0.0.1 isn't the IP adress of your pump, its your local host.
any messages in your message center?
Ace
Posted: Mon Aug 13, 2012 2:56 pm
by DR
IP addresses of components are best found in node properties, DHCP config. screen.
They should all be 192.168.0._
Ping those addresses.
Posted: Fri Aug 17, 2012 1:35 pm
by sherif4chem
yeah, you right DR and ACETO_81 .. but generally, the problem of 2707 auto-sampler which occurring many times when doing run samples sequence and suddenly the run is interrupted or stopped .. which I must make restart for all HPLC and PC and start my sequence again and sometimes the problem repeated again many times.
note: I think this problem are mainly occur to many labs having 2707 A.S waters
Posted: Fri Aug 17, 2012 3:17 pm
by DR
Any anti Virus/malware software on the computer? Does it happen at approximately the same time each day? Have you checked the PC's system and application event logs?
Posted: Sat Aug 18, 2012 8:56 am
by sherif4chem
1- there is anti virus software and updated to 2012
2- no, this happen at different times, approximately each weak may happen twice and sometimes it happen daily.
3- I checked PC's Systems ,and systems PC are good
4- Also, Firmware version I checked it up, and its OK
note : this problem is mainly occur at different labs at different locations , so I think may be this problem due to communication or defective in firmware .. Actually, am don't know