Hi All

I am currently having problems with high %CV in bracketing standards on Waters 2495 & 2489. I make up my Standard and carry out two injections from one vial position run the samples and bracket them with another two injections of the same Standard from a different vial position.

We are seeing that there is a significant difference in the areas depending on which vial position the injection was taken from for the Standards even though it is the same solution i.e the first 2 bracketing standards give lower areas that the Standards in the second position.

Has anyone experienced this problem before we use this method all the time but this seems to be a new problem. It doesnt happen all the time and seems to be intermittent. The run time for each injection is 6 minutes.

Example sample set below

Std x 2inj
Sam1 x 2inj
Sam2 x 2inj
Sam3 x 2inj
Std x 2inj

Regards
I have seen a few instances over the course of my career where the vials were the culprit in a situation such as you just described. This has always been when working with low levels of analyte. I would suggest trying a different supplier of vial, type of vial (i.e. clear instead of amber), or as a last resort, prewashing the vials a few times with your solution. Might not be the answer in your case, but it sure doesn't sound like the autosampler if you can get reproducibility from a single vial.
If the sample compartment is running below room temperature you may see this effect if you fail to equilibrate your vials for long enough.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
You can also see a vial effect if the caps are on very tight and the autosampler has to draw against a vacuum. If you can, remove the caps and see if you still have a position effect.
All standard disclaimers apply. My posts are my opinions only and do not necessarily reflect the policies of my employer.
You can also see a vial effect if the caps are on very tight and the autosampler has to draw against a vacuum. If you can, remove the caps and see if you still have a position effect.
In other words, have you overfilled the vials?
Thanks,
DR
Image
Thanks for your help Ill try this
6 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry