Ammonium adducts in ACN vs MeOH
Posted: Thu Jul 26, 2012 4:25 pm
I'm working in a lab where we're trying to get detection methods set up for a list of pesticides by LC-MS/MS. One particular class of compounds, synthetic pyrethroids (e.g. cyhalothrin, deltamethrin, permethrin, flucythrinate) was really leaving me stumped - using a gradient method recommended by Agilent with ACN/water and ammonium formate, I just couldn't see those compounds clearly by MRM or in full scans. I tried using some isocratic methods so that the full scans would be less noisy than with a gradient, and same deal.
On a whim I changed the system over to methanol/water with ammonium acetate. Using similar isocratic methods I found all of the above compounds "loud and clear" so to speak, with the strongest peak being the [M + NH4]+ adduct. I changed the aqueous phase back to ammonium formate additive while still using methanol, and got essentially the same chromatogram and spectrum.
Finally I switched back to acetonitrile/water with formate, and this time I did see the adduct, but the intensity was a good two orders of magnitude lower.
Could there perhaps be an impurity in our acetonitrile that is selectively preventing ammonium adduct formation? I can't really think of any other explanation at this point. I've never seen such a stark difference in signal between ACN and MeOH. Any thoughts from experienced folk would be welcome.
On a whim I changed the system over to methanol/water with ammonium acetate. Using similar isocratic methods I found all of the above compounds "loud and clear" so to speak, with the strongest peak being the [M + NH4]+ adduct. I changed the aqueous phase back to ammonium formate additive while still using methanol, and got essentially the same chromatogram and spectrum.
Finally I switched back to acetonitrile/water with formate, and this time I did see the adduct, but the intensity was a good two orders of magnitude lower.
Could there perhaps be an impurity in our acetonitrile that is selectively preventing ammonium adduct formation? I can't really think of any other explanation at this point. I've never seen such a stark difference in signal between ACN and MeOH. Any thoughts from experienced folk would be welcome.