Peaks coming out in the next run, in gradient mode
Posted: Thu Jul 26, 2012 9:35 am
Oops, I'm the world's worst chromatographer.
For ages I've had a variety of typical flow-rates and scout gradients using a 100*2mm 3u C18 column which I might use as a starting point on a novel analyte. The last couple of days I've been using a fairly wide gradient, 5-95% acetonitrile (versus 0.1% formic acid; this is LC-MS) over 20 min at 280uL/min, followed by 5 min at 95% acetonitrile because some of my analytes seemed to be strongly retained and I wanted to make sure they were really gone.
The peaks are coming out in the run after the one where they were injected. I know this happens in isocratic chromatography, but I haven't seen it in gradient chromatography until now. The strange part is that they don't come out particularly late; they're eluting at about 11min into the run, i.e. 10 min into the gradient given the dead-volume on the system. This definitely isn't autosampler needle carry over, as the peaks don't appear in the run where they were injected, only the subsequent run (there are no peaks in a further blank run; if I inject (1)blank; (2)sample; (3)blank; (4)blank, the only run with the expected peaks is (3)).
I always assumed that analytes were more-or-less stationary on the stationary phase until the gradient reached a percentage where they begin to partition into the mobile phase; then there's a period where they're partitioning and moving at less than full mobile-phase speed (separation is now behaving like an isocratic run) but as the gradient continues, they get to a stage where they're partitioned entirely in the mobile phase. As a result, I'd expect a compound that is capable of moving at all at only half-way up my gradient to be legging it along the column at full speed when running at 95% acetonitrile, so I can't see how on earth these peaks were retained for 5 minutes at 95%, but then able to leave the column at only 11min = 50% in the next run.
My only possible explanation is that they are sticking to the column by two reasons; hydrophobic interactions and something completely different. A relatively low percentage of acetonitrile is overcoming the hydrophobic interactions, but the other interaction is constant, so my analytes never partition fully off the stationary phase; instead they spend the greater part of the gradient behaving in an isocratic way. I don't know if this is possible? The compounds are nod-factors (http://en.wikipedia.org/wiki/Nod_factor) so they are very multifunctional; they've got a sulphate group, a long fatty chain, and a row of glucosamine sugars. I don't really believe my explanation because I'd expect the long fatty chain to act as a hydrophobic anchor and need quite a high percentage of acetonitrile to elute, and there's no particular reason why a sulphate should stick to a Luna C18 column, or a glucosamine for that matter, is there??
Incidentally, I don't think I'm alone in finding these things weird. Looking at literature, the solvbent mixes that people seem to need to elute them vary by about 50% organic between references using very similar columns, which worries me.
Any explanations/comments gratefully received...
For ages I've had a variety of typical flow-rates and scout gradients using a 100*2mm 3u C18 column which I might use as a starting point on a novel analyte. The last couple of days I've been using a fairly wide gradient, 5-95% acetonitrile (versus 0.1% formic acid; this is LC-MS) over 20 min at 280uL/min, followed by 5 min at 95% acetonitrile because some of my analytes seemed to be strongly retained and I wanted to make sure they were really gone.
The peaks are coming out in the run after the one where they were injected. I know this happens in isocratic chromatography, but I haven't seen it in gradient chromatography until now. The strange part is that they don't come out particularly late; they're eluting at about 11min into the run, i.e. 10 min into the gradient given the dead-volume on the system. This definitely isn't autosampler needle carry over, as the peaks don't appear in the run where they were injected, only the subsequent run (there are no peaks in a further blank run; if I inject (1)blank; (2)sample; (3)blank; (4)blank, the only run with the expected peaks is (3)).
I always assumed that analytes were more-or-less stationary on the stationary phase until the gradient reached a percentage where they begin to partition into the mobile phase; then there's a period where they're partitioning and moving at less than full mobile-phase speed (separation is now behaving like an isocratic run) but as the gradient continues, they get to a stage where they're partitioned entirely in the mobile phase. As a result, I'd expect a compound that is capable of moving at all at only half-way up my gradient to be legging it along the column at full speed when running at 95% acetonitrile, so I can't see how on earth these peaks were retained for 5 minutes at 95%, but then able to leave the column at only 11min = 50% in the next run.
My only possible explanation is that they are sticking to the column by two reasons; hydrophobic interactions and something completely different. A relatively low percentage of acetonitrile is overcoming the hydrophobic interactions, but the other interaction is constant, so my analytes never partition fully off the stationary phase; instead they spend the greater part of the gradient behaving in an isocratic way. I don't know if this is possible? The compounds are nod-factors (http://en.wikipedia.org/wiki/Nod_factor) so they are very multifunctional; they've got a sulphate group, a long fatty chain, and a row of glucosamine sugars. I don't really believe my explanation because I'd expect the long fatty chain to act as a hydrophobic anchor and need quite a high percentage of acetonitrile to elute, and there's no particular reason why a sulphate should stick to a Luna C18 column, or a glucosamine for that matter, is there??
Incidentally, I don't think I'm alone in finding these things weird. Looking at literature, the solvbent mixes that people seem to need to elute them vary by about 50% organic between references using very similar columns, which worries me.
Any explanations/comments gratefully received...