%RSD for injection repetability
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
%RSD for 6 injections of API(in 100% acn) is about 1.7% if injected from the same vial , but is less than 0.5% for 6 injections from different vials. the autosampler temp is 5C so there is very little chance of evaporation. Is there an explanation for this senario?
-
- Posts: 40
- Joined: Tue Apr 24, 2012 3:36 pm
is this just from one set, it is easily within the realm of possibilty to get those results given a small enough data size.
-
- Posts: 72
- Joined: Tue Jan 11, 2011 6:41 pm
I would try that using self sealing septa.
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
What injection volume are you removing?
Could you be creating a vacuum effect in the vial which could affect the ability of the autosampler syringe to draw samples.
A vacuum could be created if you overtighten cap and septa at room temperature then cool to 5C which could affect injection performance.
Could you be creating a vacuum effect in the vial which could affect the ability of the autosampler syringe to draw samples.
A vacuum could be created if you overtighten cap and septa at room temperature then cool to 5C which could affect injection performance.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
Can you see an increasing or decreasing trend with your injections from 1 vial?
Or is it just random?
Ace
Or is it just random?
Ace
-
- Posts: 1888
- Joined: Fri Aug 08, 2008 11:54 am
If the difference is reproducible time-after-time in successive experiments, ignore part 1 of this post!
Part 1:
coincidentally, see Veronika Meyer in LC-GC this month (August 2012) for a rather long investigation of the errors in experimental measuerments of errors. If you've measured the standard deviation of the mean in one experiment of n replicates, you could go on to carry out m experiments, and get m different standard deviations. Then you could calculate the standard deviation of the standard deviation. Meyer's point is that it's often quite big relative to the standard deviation of the mean. If you found your difference in only a couple of experiments, it's quite possible it's down to chance.
To be honest I can't imagine many scenarios where even as much as a 2-fold random error in the standard deviation of the mean of analytical replicates really matters.
Part 2:
I agree with aceto_81 that ongoing trends are more worrying than obviously random fluctuations; JGK's right about vacuums. Another thing I've seen happen in a cooled autosampler is that a drop of condensation can appear in the dimple left at the top of a vial after a needle's gone through the septum. If you're unlucky, it then dilutes the next injection, and so on...
Part 1:
coincidentally, see Veronika Meyer in LC-GC this month (August 2012) for a rather long investigation of the errors in experimental measuerments of errors. If you've measured the standard deviation of the mean in one experiment of n replicates, you could go on to carry out m experiments, and get m different standard deviations. Then you could calculate the standard deviation of the standard deviation. Meyer's point is that it's often quite big relative to the standard deviation of the mean. If you found your difference in only a couple of experiments, it's quite possible it's down to chance.
To be honest I can't imagine many scenarios where even as much as a 2-fold random error in the standard deviation of the mean of analytical replicates really matters.
Part 2:
I agree with aceto_81 that ongoing trends are more worrying than obviously random fluctuations; JGK's right about vacuums. Another thing I've seen happen in a cooled autosampler is that a drop of condensation can appear in the dimple left at the top of a vial after a needle's gone through the septum. If you're unlucky, it then dilutes the next injection, and so on...
6 posts
Page 1 of 1
Who is online
In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: John Guajardo and 26 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: John Guajardo and 26 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science