LG05 » Wed Jul 25, 2012 6:16 pm
I'm setting up a new HPLC and ran into some issues when I tried to equilibrate a brand new GL Sciences NH2 (150 x 0.3 mm, 3um) column. When I have the inline filter, guard column, and column connected I get relatively high pressures (~1800 psi), the inline filter leaks, and there is little to no flow coming off the column. I tried tightening the filter and the fittings connecting it but it made no difference. Then I removed the column and checked the flow through just the filter and guard column. In that case, the pressure is only ~50psi and the flow is normal.
I haven't run any samples yet. This has just been using 100% IPA to equilibrate the column. But I'm thinking there's either air or particles clogging the column?? Any ideas on how to remedy this situation? Would reversing the column and flushing it out work even if it means having a major leak from the inline filter??