waves in the baseline of TCD detector
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
Hello, I'm using Thermo GC Ultra and in the baseline there is some sinwave. Before one mounts I replased the reduction valve of the argon background gas bottle and solve the problem but now it appear again. The valve is SIAD for laboratory usage. The pressuse in the bottle is 200 bars and 6,5 after the valve. We open the close the bottle every time we use the device. We don't touch the regulator of the valve.
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
remofil,
This does not solve your problem (sorry) but I would advise against turning off the flow to the TCD unless it will be sitting idle for weeks at a time.
You might look at lowering the temperature of the TCD to see if this changes the characteristics of the sine wave. You might have a heater problem.
Best regards,
AICMM
This does not solve your problem (sorry) but I would advise against turning off the flow to the TCD unless it will be sitting idle for weeks at a time.
You might look at lowering the temperature of the TCD to see if this changes the characteristics of the sine wave. You might have a heater problem.
Best regards,
AICMM
-
- Posts: 13
- Joined: Thu Mar 08, 2012 11:59 pm
Hi remofil,
I recently had the same problem. Is the vent to your TCD inside the oven?
In my case, we had the TCD vent inside the oven, and every time the oven door was closed (i.e., we had the GC running or stablizing), the baseline was just horrible - it looked like a sine wave. When we opened the GC door, the baseline stabilized. We realized the noise was coming from the TCD vent, so we just rerouted it to the outside of the oven, and our baseline problem was solved.
Hope this simple solution works for you!
GHGirl
I recently had the same problem. Is the vent to your TCD inside the oven?
In my case, we had the TCD vent inside the oven, and every time the oven door was closed (i.e., we had the GC running or stablizing), the baseline was just horrible - it looked like a sine wave. When we opened the GC door, the baseline stabilized. We realized the noise was coming from the TCD vent, so we just rerouted it to the outside of the oven, and our baseline problem was solved.
Hope this simple solution works for you!
GHGirl
-
- Posts: 102
- Joined: Mon Jan 10, 2011 1:38 pm
TCD is conductivity dependent, so it's a temperature or a flow effect you are seeing.
Temperature control can be checked by turning off the heater and see what happens with the "wave".
Flow effects: turn of carrier gas flow. Now you only have make up gas. If still there it's the make up gas regulation or the reference gas regulation.
jaap de zeeuw, Restek Corporation
Temperature control can be checked by turning off the heater and see what happens with the "wave".
Flow effects: turn of carrier gas flow. Now you only have make up gas. If still there it's the make up gas regulation or the reference gas regulation.
jaap de zeeuw, Restek Corporation
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science