Alera » Mon Aug 23, 2010 3:13 pm
MestizoJoe, look at it from the efficiency and ease of prep point of view. After you collect your fractions (about 5 mLs), you have to centri-vap them, which takes time and increases salt concentration significantly (potential issues with proteins which might denature at some point).
Using a membrane unit, you load all 5 mLs into the unit and spin it, concentrating and removing salt at the same time (salt concentration never increases). Once the volume is reduced sufficiently, you add your final buffer to dilute the sample and spin again, so if you spin twice, say 5mL to 0.5mLs, you will have only 1% of the salt left in the sample.
The reason I preferred this method over SEC was that it provides for less sample transfers (and therefore, losses), as sample concentration and buffer exchange occurs at the same time.
I used this procedure for fairly small to moderate (similar to your scaled up) sample amounts, but it sounds in your "scaled up" prep it should be fesible as well.