%Methyl Ester Content & Split Splitless Help
Posted: Thu Jul 19, 2012 9:49 pm
I have inherited an HP 5890 w/split splitless no EPC, FID. With a restek Stabiliwax 30m, 1micron film,
0.53mm ID column. I don't think it has been used since purchased many years ago.
First time working with split/splitless inlet. Not exactly sure what I am doing. Method requires 2mL/min carrier gas, 100mL split on a 30m 0.25mm ID column with 0.1-0.25 micron film thickness.
I have been testing at 2-4mL/min on the column, 100 mL off the split vent. The purge vent is a mystery to me, it looks plugged, but I can get flow out of it when I turn the purge inlet valve ON. Don't know how much flow I am supposed to be getting of this vent or if the valve should even be open? I have been running with it OFF. When I turn the valve ON the flow through the column goes up 5mL/min?
Anyway have been injecting toluene solvent blank and 100mg FAME dissolved in 10mL toluene as the method requires. All i get is one huge tailing solvent peak and no ester peaks. I imagine I am either getting too much sample on the column or I am losing sample on the inlet and not getting it on the column. The liner is straight tapered no wool. There is a new o-ring on it, but the liner has some small chips on the ground glass upper surface. I have the column seated 1-2 mm above the taper of the liner. I have about a half meter of 0.53m guard column connected with a glass press tight connection (came with another phenomenex column). I have checked all column connections/FID jet. One last thing is the "gold" seal at the bottom of the inlet, this one isn't gold but blue oxidized and sits on a metal washer.
Any suggestions or help would be greatly appreciated. Thanks.
0.53mm ID column. I don't think it has been used since purchased many years ago.
First time working with split/splitless inlet. Not exactly sure what I am doing. Method requires 2mL/min carrier gas, 100mL split on a 30m 0.25mm ID column with 0.1-0.25 micron film thickness.
I have been testing at 2-4mL/min on the column, 100 mL off the split vent. The purge vent is a mystery to me, it looks plugged, but I can get flow out of it when I turn the purge inlet valve ON. Don't know how much flow I am supposed to be getting of this vent or if the valve should even be open? I have been running with it OFF. When I turn the valve ON the flow through the column goes up 5mL/min?
Anyway have been injecting toluene solvent blank and 100mg FAME dissolved in 10mL toluene as the method requires. All i get is one huge tailing solvent peak and no ester peaks. I imagine I am either getting too much sample on the column or I am losing sample on the inlet and not getting it on the column. The liner is straight tapered no wool. There is a new o-ring on it, but the liner has some small chips on the ground glass upper surface. I have the column seated 1-2 mm above the taper of the liner. I have about a half meter of 0.53m guard column connected with a glass press tight connection (came with another phenomenex column). I have checked all column connections/FID jet. One last thing is the "gold" seal at the bottom of the inlet, this one isn't gold but blue oxidized and sits on a metal washer.
Any suggestions or help would be greatly appreciated. Thanks.