DL Alpha Tocopherol Acetate by GC

Posted: Fri Jun 03, 2005 8:45 am
by Jane Hing
I've been trying to run a BP method for tocopherol acetate by GC and although it looks good and the resolution is great I have two peaks for my active :roll: . The Natural Tocopherol under the same conditions produces only one peak.

Is this normal? I'm using a packed column.
:roll:

Posted: Fri Jun 03, 2005 12:36 pm
by BG
What is the temperature of your injection port because all the applications i can see, put the injector at 300°C (with a retention gap before the analytical column ).
Maybe there is a bad vaporisation and you fragment your band width ....

Posted: Fri Jun 03, 2005 2:46 pm
by Consumer Products Guy
The peaks elute close to each other, correct? The first peak is likely to be the tocopherol and the second peak the tocopheryl acetate (they have different GCMS spectra as well). We normally do the tocopheryl acetate assay by HPLC on a C8 column, with 95% acetonitrile.