- Posts: 45
- Joined: Fri Feb 05, 2010 8:59 am
I'm working with GCMS from several years and on the last two years I observe into the chromatogram four ghost peaks WITHOUT injection, only run : octamethyl cyclo tetra siloxane, decamethyl cyclo penta siloxane, dodecamethyl cyclo hexa siloxane and tetra deca methyl epta siloxane. My instrument is Shimadzu GC-QP2010 and the tested columns with identical results was ZB5MS, ZB5Semivolatiles (Zebron) DB5MS-UI (Agilent), all 60m , 0,25 mm I.D and 0.25 u as film thickness. All columns where conditionated as recommended producer protocol.
My first test to understand the problem was to verify possible contamination coming from GC and I have tried to bypass entirely the injection system by connecting the column with external pressure regulator with no positive result.
The second step was to change helium source (Helium BIP) with three other different gas and no result.
The third step was to replace the in-line triple filter and no result again.
Now, the peaks are quite high (with 1 kV of EM the peak hight could raise 800000).
The first peak is very broad, the second is quite better and the last two are well focalized.
The eluting temperature is around 100°C for the first peak and 200-210 for the last.
I can suppose that the contamination come from the column itself due to heat ramp but seems to me very strange considerating the importance of columns producers.
Any idea ?
Regards,
Robertino Barcarolo, Italy