Agilent Chemstation Calibration Curve
Posted: Tue Jul 17, 2012 6:41 pm
I am working with Agilent Chemstation and am trying to analyze for 3 compounds. I have created an 8 point calibration curve and have a good linear regression for each compound. My calibration curve ranges from 0.05 ug/ml to 10.0 ug/ml. When I go to reprocess my calibration curve for 0.05 and 1.0 ug/ml I receive amounts such as 0.00 for the first compound, 2.3e-3 for the second and 3.4e-1 for the third compound at standard 0.05 and for 1.0 ug/ml I receive 1.1e-3 for the first compound 2.4e-2 for the second and 3.4e-1. At first I thought maybe I should recalibrate or make a new curve but when I do the first two cal points (0.05 and 1.0) and then go back to the first cal-point 0.05 I get the results (5.5e-2 5.4e-2) that I am looking for. Another frustraiting issue is that samples that are in the middle of my cal-range (1.0 ug/ml) come as expected. Also when I go to calculate other samples at 0.10 concentration the area is the same as my standard in the calibration curve (44.372) but the concentration amounts are not the same. Any clues as to how I might be able to fix this?