Hey, when I run a sample set and I won't be at work when it's done, I always include a shutdown/wash method (chosen when you press start). This method turns off the detector, washing the column, then shuts off flow.
When I come back to the system, communication is lost with the system. I can't control anything from the run sample window. I have to go into configuration, take my system offline and then back online and everything works.

One other related question; say I want to stop a sample set that's running, it automatically switches to the shutdown method I selected, is there any way to stop that from happening? Generally if you have to interrupt a sample set and stop it, it means you need to fix something and run it again...

Thanks!
Instead of specifying a shutdwon method when starting a sample set method, I leave shutdown method blank and start a second sample set method that conditions the column w/ the preferred cleanout method and shuts the lamp off. The shutdown method after this sample set method keeps things off (after running a little more cleanout).
Thanks,
DR
Image
When I come back to the system, communication is lost with the system. I can't control anything from the run sample window. I have to go into configuration, take my system offline and then back online and everything works.
It seems there may be an admin function to take system off line when sample set is complete? Does it happen for every run?
One other related question; say I want to stop a sample set that's running, it automatically switches to the shutdown method I selected, is there any way to stop that from happening? Generally if you have to interrupt a sample set and stop it, it means you need to fix something and run it again...
You can alter the sample set, remove the shutdown, save and continue the sample set, prior to aborting the run.
This is a known issue with FR5 and agilent equipment. The way that we got around it is to create a shutdown method set and add it as the last line of teh sample set.
4 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry