DL Alpha Tocopherol Acetate by GC

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I've been trying to run a BP method for tocopherol acetate by GC and although it looks good and the resolution is great I have two peaks for my active :roll: . The Natural Tocopherol under the same conditions produces only one peak.

Is this normal? I'm using a packed column.
:roll:

What is the temperature of your injection port because all the applications i can see, put the injector at 300°C (with a retention gap before the analytical column ).
Maybe there is a bad vaporisation and you fragment your band width ....

The peaks elute close to each other, correct? The first peak is likely to be the tocopherol and the second peak the tocopheryl acetate (they have different GCMS spectra as well). We normally do the tocopheryl acetate assay by HPLC on a C8 column, with 95% acetonitrile.
3 posts Page 1 of 1

Who is online

In total there are 30 users online :: 0 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 30 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry